[molecular biology work] ◈_gel_filtration_protocol-1

This topic has 0 replies, 1 voice, and was last updated 16 years, 6 months ago by Seung Jae Jeong.

  • Author
    Posts
    • #18884

      Seung Jae Jeong
      Participant

      Gel filtration

       

      Kim Jinyoung

       

      1.     p43,p38,wild type MEF cell 100π dish 20장에 culture(37 CO2 5%).

      2.     TE처리후 cold 1XPBS spindown 하고, 다시 한번 cold 1XPBS washing한다.

      3.     hypotonic solution(10mM HEPES pH7.6, 10mM KCL, 1.5mMMgCl2, 0.5mM EGTA, 10mM NaF, 1mM PMSF, Protease  inhibitor cocktail ) syringe(needle size 23G) lysis 시킴.

      4.     4, 14000 rpm에서 15min centrifuge.

      5.     transfer sup. into new tube, 4, 14000 rpm에서 15min centrifuge

      6.     sup과 동일 volume solution(10mM HEPES pH7.6, 150mM KCL, 1.5mMMgCl2, 0.5mM EGTA, 10mM NaF, 1mM PMSF, Protease  inhibitor cocktail )을 첨가해서 80mM KCl solution으로 만든다.

      7.     8mg/ml vivaspin(VIVASCIENCE)을 이용하여 농축 시킨다(4, maximum 4000 g Tommy High speed). 0.22um filter filtering을 한 후 단백질 농도를 잰다.

      8.     (10mM HEPES pH7.6, 80mM KCL, 1.5mMMgCl2, 0.5mM EGTA, 10mM NaF, 1mM PMSF, Protease  inhibitor cocktail ) buffer FPLC pump column(Pharmacia, Sephacryl S-300 High resolution, separate range 10Kda-1500KDa ) equilibrium 시킨다.

       

      l  FPLC를 사용하기 전에는 꼭 pump calibration pump washing을 꼭 해준다.

      Pump calibration FPLC protocol 107page, pump washing 97page에 잘 설명되어 있음.

                protocol 마지막 부분 참고!!

       

       

      9.     sample 500ul(4mg/500ul) injection flow rate 1.0ml/min, fraction size 1ml gel filtration program 4℃에서  돌린다.

                Program

                Set pressure limit : 0.35MPa

                Set flow rate : 1.0 ml/min

                Set fraction size : 1.0ml

                Set equilibration volume : 1.0ml

                Set sample injection volume : 0.5ml

                Set elution volume : 180ml

                Set wash volume : 0ml

      10.  일부 fraction을 취해서 western blotting을 한다.

               
      70ul loading

      8.12 Removing trapped air bubbles

      Remove trapped air bubbles in the flow path by purging the pump with liquids. Use the liquids in the following order: 1. deionized water, 2. 20% ethanol, 3. deionized water and 4. buffer solution.

      Note: All liquids used must be degassed.

      Note: When using degassed ethanol, make sure that the concentration dose not fall below the required value.

      Puring can be done manually through inlet A1, while carefully immersing the tubing in the respective liquid. Set the infection valve to position WASTE. Run 30 ml of each liquid at 50 ml/min. press the pause/cont button to start and stop the pump when changing liquid.

      An automatic purging procedure that uses additional inlet tubings is decribed below:

      1  Connect at inlet tybing to buffer valve port 4. Immerse the tubing in a vessel filled with deionized water.

      2  Connect an inlet tubing to buffer valve port 3. Immerse the tubing in a vessel filled with 20% ethanol.

      3  Connect an inlet tubing to buffer valve port 2. Immerse the tubing in a vessel filled with deionized water.

      4  Connect an inlet tubing to buffer valve port 1 is immersed in a buffer solution.

      5  Fill the empty inlet tubing manually.

      6  In Templates, select the System Wash Method under Application template.

      7  Select ports 2, 3 and 4 to be washed. Port 1 is pre-selected. De-select port B.

      8  Press OK to start the run.

         The pump first draws the liquids in the following order: deionized water, ethanol, deionized water, and finally the buffer solution .
         Consequently, when the run is finished, the system is filled with the buffer solution connected to buffer valve port 1.